Yes, and I did. I guess I was not clear enough. I placed the Fieldpiece on a stand and took the reading. I then took a second reading from the exact same spot and they were not the same. I then tried about three different locations each at a slightly different distance from the reflective tape on the platter, and at different angles and these readings also did not match. They varied +/- 1 RPM, ie 33.845, 34.352, 33.908 etc. Each time the device was fixed and never moved. I had the KAB strobe on the platter and it did not waver, so I assume the platter was not changing speed. The stylus was not on a record and I did not adjust the speed on the motor controller.
For such a device to be useful, I would think that any reading from any distance within say 2-5 feet from the platter should be extremely close if not exactly the same. The point is that without getting consistent readings, there is no way of telling exactly how fast the platter is spinning. My conclusion is that for me, this device failed the test.
If anyone would like to buy this mint condition device for perhaps a different use, please contact me.
For such a device to be useful, I would think that any reading from any distance within say 2-5 feet from the platter should be extremely close if not exactly the same. The point is that without getting consistent readings, there is no way of telling exactly how fast the platter is spinning. My conclusion is that for me, this device failed the test.
If anyone would like to buy this mint condition device for perhaps a different use, please contact me.