500mV (0dB) on the voltmeter?


I am tuning up my Nakamichi CR-3A cassette deck and am trying to understand the meaning of a dB value that follows a mV value. In relation to measuring output with a 400Hz input, the service manual refers to obtaining "500mV -2dB on the AC voltmeter" from the output jacks. In another spot it refers to obtaining "500mV (0dB) on the AC voltmeter" with a 19kHz input. I am using a Brymen 867 multimeter. I can measure 500mV from the output jacks easily enough but am left with two questions: 1) what do these dB values refer to? 2) is there any difference between the dB value within or without parentheses?
brymul
I love this.

I tell my track and field athletes

"Math is hard, so is track, deal with it"

Thanks Al. Can’t understand why they would "arbitrarily" refer to something in the service manual. Also, why not just refer directly to 397 mV?
Can’t understand why they would "arbitrarily" refer to something in the service manual.
In the context of voltage a given number of dB refers to the ratio between two voltages, as I indicated. Therefore "0 dB" must be defined as corresponding to **some** specific voltage. And there is not a single specific voltage that can be defined which will be best suited for all applications. So in that sense whatever voltage 0 dB is assigned to in a particular case will have been chosen "somewhat arbitrarily" (as I put it in my previous post), meaning that it will have been chosen by a particular organization or individual for a particular purpose.

In some contexts 0 dB may be defined as representing 1 volt, in which case voltages may be stated as dBV, the "V" signifying that the number of db is relative to 1 volt. "dBu" is also often used, 0 dBu corresponding to about 0.775 volts. In this case they chose 0 dB as corresponding to 0.5 volts, probably for reasons related to the signal levels used on the cassette tapes.

Also, why not just refer directly to 397 mV?
Some voltmeters provide a scale calibrated in dB, in addition to a scale calibrated in voltage. Especially older analog meters, if I recall correctly. Also, VU meters on tape decks are usually calibrated in dB, so specifying input voltages in dB may make it more convenient to correlate and compare changes in input voltage to changes in VU meter readings.

Regards,
-- Al