Hi Bryon,
I looked through the Fluke 115 literature. The one notable thing I found is that I think you should perform the measurement using the "AC Volts" mode, not the "AC Millivolts" mode. The meter's input is AC-coupled in the "AC Volts" mode, meaning that any DC that may be present (which might otherwise confuse the measurement) will be blocked in that mode. "AC Millivolts" is DC-coupled, which is undesirable for what you are trying to do. The "AC Volts" mode can resolve down to 1 mv, which should be more than adequate.
On the Model 115 the "AC Volts" mode corresponds to the large rotary mode select switch being at the position that is immediately to the right of the "Off" position.
Also, given my uncertainty about the need for connecting the ground wire to an outlet screw, try it both with and without that wire, if necessary and if you already haven't.
Hope that helps. Best,
-- Al
I looked through the Fluke 115 literature. The one notable thing I found is that I think you should perform the measurement using the "AC Volts" mode, not the "AC Millivolts" mode. The meter's input is AC-coupled in the "AC Volts" mode, meaning that any DC that may be present (which might otherwise confuse the measurement) will be blocked in that mode. "AC Millivolts" is DC-coupled, which is undesirable for what you are trying to do. The "AC Volts" mode can resolve down to 1 mv, which should be more than adequate.
On the Model 115 the "AC Volts" mode corresponds to the large rotary mode select switch being at the position that is immediately to the right of the "Off" position.
Also, given my uncertainty about the need for connecting the ground wire to an outlet screw, try it both with and without that wire, if necessary and if you already haven't.
Hope that helps. Best,
-- Al