The book addresses Degas as follows:
XIV.2.1 Dissolved Gases & Degas: Fluids exposed to air will absorb air and the solubility of air in water is inversely proportional with temperature and can decrease by about 25% between 20°C/68°F and 40°C/104°F. Air that is dissolved in the fluid will interfere with cavitation. As the cavitation bubble forms, the dissolved air in the fluid migrates into the cavitation bubble preventing maximum cavitation intensity when the cavitation bubble collapses. The degas process operates the ultrasonics with a tank of fresh liquid to remove some or most of the dissolved air. Degas efficiency is dependent on frequency, power and volume. Low frequency 40-kHz and less can take 30 minutes or more to fully degas a large tank, while 80 kHz and greater can accomplish degas in as little as 5 minutes. During the degas process, bubbles may be seen rising in the fluid; the fluid may go from cloudy to clear; there may be a change in sound; and the surface can change to smooth with just a slight rippling effect caused by the ultrasonics.
The book does not state to degas each day if using the same bath although it is implied with the first sentence above. If the book is ever revised, XIV.2.1 will be revised to be more explicit.