Radio Shack SPL Meter mods corrections


I know that others here were interested in the specs to correct the non-linearities of the factory stock RS SPL meter. The figures below are to be either added to (+) or subtracted from (-) the readings that you obtain from your factory stock meter. In other words, the meter in stock form is EXTREMELY deficient when it comes to taking low frequency measurements. Keep in mind that these corrections are only valid under the following conditions. The meter must be set to C weighting, using 1/3 octave pink noise (easily available from various CDs), with the mic pointed at the speaker. These measurements were verified on both the RS analogue and digital meters using laboratory grade test equipment. Baseline testing was done using the 80 dB scale for reference purposes. While on the low side, this should give you a good baseline as to what your actually getting out of your system. Needless to say, if you had BIG peaks in the bass region with the stock meter and you weren't calculating in these correction figures, you're in even worse shape than you thought.

10Hz +20.5
12.5Hz +16.5
16Hz +11.5
20Hz +7.5
25Hz +5
31.5Hz +3
40Hz +2.5
50Hz +1.5
63Hz +1.5
80Hz +1.5
100Hz +2
125Hz +0.5
160Hz -0.5
200Hz -0.5
250Hz +0.5
315Hz -0.5
400Hz 0
500Hz -0.5
630Hz 0
800Hz 0
1KHz 0
1.25Khz 0
1.6KHz -0.5
2Khz -1.5
2.5Khz -1.5
3.15Khz -1.5
4KHz -2
5KHz -2
6.3KHz -2
8KHz -2
10Khz -1
12.5KHz +0.5
16KHz 0
20KHz +1

In order to correct the low frequency roll-off, you can do the following modifications to your meter. This will make the meter FAR more sensitive to low frequencies and allow measurements with very good accuracy to well below 20 Hz. Due to the increased sensitivity at very low frequencies, it is possible for low frequency "thumps" to slam the meter if using a very low SPL setting. In order to prevent meter movement damage, take precaution not to peg the meter off scale on a regular basis. If this is occuring regularly, you either need to move up to a higher SPL range on the meter or take more caution as to how your performing your tests.

All of the following capacitors must be rated for AT LEAST 15 volts or so. Size does matter, so try to use the smallest package possible.
C1 & C2 are changed from 1 uF to 10 uF
C3 & C4 are changed from 1 uF to 47 uF
C7* is changed from 10 uF to 220 uF
C8 is changed from 100 uF to 470 uF
C9 is changed from 22 uF to 220 uF
C15 is changed from 100 uF 220 uF

While all of the above parts are spec'd as microfarads, the following is in picofarads. Do not confuse the two values or the meter will not work very well at all. This last change helps minimize high frequency roll-off that is inherit in the stock microphone

C12 is changed from 33 pF to 12 pF.

Please note that all of the above parts can be soldered directly in place of the originals EXCEPT C7. Due to its location, a "normal sized" 220 uF cap will be too big to allow the case to close correctly. In order to get around this, simply solder it on the "solder side" of the board instead of on the "parts side" of the board. Pay special attention to the POLARITY of ALL of the caps as you pull them out to replace them. For this reason, i recommend pulling and replacing the caps one at a time to minimize confusion. Once all of these mods are done, the meter is more than accurate enough for anything that a home audio enthusiast would ever need use of. If you really want to "get crazy", you can remove the factory installed mic and either remotely mount it on a "wand" or make use of a calibrated mic like those available from Old Colony.

I hope this helps some of you out and sorry it took so long to dig all of this up. Sean
>
sean
Sean, to offer a "thank you" seems unworthy. I have had the privelage of reading countless informative posts during my stint on Audiogon. This has to be the most instructive, informative, well researched post I have had the pleasure to read. You certainly deserve something above a commendation. THANK YOU! Joe.
That was Awesome! I did a test using my new Terrasonde The Audio Toolbox, albeit in nowhere near lab conditions, nor do I suspect the audio toolbox qualifies as lab grade equipment, and my data roughly agrees with your data above.

I thought that some of the bass roloff had to do with the C weighting of the meter, do you have any insight in to that?

Niels.
Folks, i'd like to take credit for all of the info presented above, but i can't. I simply combined information that i've picked up from various sources and presented it together as a package for convenience sake. Some of the websites that i garnered all of this "well documented" info from are no longer operating. Since that is the case, i can't forward you to some of them. If your truly interested in more background on the data presented here, email me directly and i'll forward what i have. Take my word for it though, all of it presented here has been well tested and is quite accurate when used in the manner specified. Keep in mind that when comparing readings between one of these SPL meters and a computer program, you are at the mercy of the sound card and microphone used for the puter. As such, most "sound blaster" type cards along with all of the "generic" versions are HORRIBLY non-linear. Combine this with a non-calibrated microphone and you would be COMPOUNDING the results, giving you nothing but error filled data at best. Glad you "guys" ( and gals too.... ) liked this and can put it to use. Sean
>

I agree with the above posts re: this excellent thread. Thanks Sean, for the time and effort it took to put your correction table together for the RS SPL meter. I copied the thread. Thanks, Craig.
Not to be dense Sean, but these adjustments are to be factored in when using the RS SPLM as described in Audiophilia? http://www.audiophilia.com/hardware/spl.htm Thanks for the tip, but I am very afraid, that as you said, I'm now in worse shape than I thought!